Skip to content
  • MySensors
  • OpenHardware.io
  • Categories
  • Recent
  • Tags
  • Popular
Skins
  • Light
  • Brite
  • Cerulean
  • Cosmo
  • Flatly
  • Journal
  • Litera
  • Lumen
  • Lux
  • Materia
  • Minty
  • Morph
  • Pulse
  • Sandstone
  • Simplex
  • Sketchy
  • Spacelab
  • United
  • Yeti
  • Zephyr
  • Dark
  • Cyborg
  • Darkly
  • Quartz
  • Slate
  • Solar
  • Superhero
  • Vapor

  • Default (No Skin)
  • No Skin
Collapse
Brand Logo
  1. Home
  2. Announcements
  3. Sensebender Micro

Sensebender Micro

Scheduled Pinned Locked Moved Announcements
584 Posts 84 Posters 401.8k Views 35 Watching
  • Oldest to Newest
  • Newest to Oldest
  • Most Votes
Reply
  • Reply as topic
Log in to reply
This topic has been deleted. Only users with topic management privileges can see it.
  • nivocN nivoc

    @tbowmo Thanks. No worries the electronic won't touch the salt or water. The salt/water is just within the same closed container to bring the enclosed air to exactly 75.3% rel hum (at 25C).

    I followed this howto: But there are many other similar ones.
    http://www.kingofthehouse.com/hygrometer/

    DwaltD Offline
    DwaltD Offline
    Dwalt
    wrote on last edited by
    #353

    @nivoc I setup several sensebenders on battery last summer and noticed they were a few % below two dht22 I had been running for months. I did the saturated salt test on three sensebenders with one dht22 and one mechanical hygrometer and the sensebenders all read 74-75% while the DHT and hygrometer read 78-80%. Not a perfect test and i dont know which devices were correct but one thing I noticed was the sensebenders consistently provided the same result across different devices. Relative humidity is relative.

    Veralite UI5 :: IBoard Ethernet GW :: MyS 1.5

    1 Reply Last reply
    0
    • nivocN Offline
      nivocN Offline
      nivoc
      wrote on last edited by
      #354

      @Dwalt
      Thanks for the info. I'm currently opening the box every several hours to test the salt soup. My room has something between 50 and 60 rel. (so below 75). And every time I open the box this (see pic). It drops and goes back up to 82 or 83. Than it stabilizes there over the next couple hours. There are two sensbenders in the box - thats what the two line are. And yes they are remarkable equal. But it always goes to 82 - i would hope for 75. But I will repeat this procedure until it starts to stabilizes after every opening a little bit below the value that was there before the opening. Because if that happens i am below the dry point and than I will test in a room with humidity higher than 80 and test if soup can do the same into the other direction and holds it stable for many openings. THAN i know that the salt is performing is job correctly and know that the test is good and the senbender is wrong :-) or hopefully I learn the opposite.

      upload-555b12a4-6bc0-4e07-acf7-976c4122e6d6

      1 Reply Last reply
      0
      • nivocN Offline
        nivocN Offline
        nivoc
        wrote on last edited by
        #355

        Yes it looks like that the sensors are 8% off. I tested the other way around >opening at ~100rel and closing and same + exactly back to 82/83 and it should be 75% :-).

        upload-e6c04b83-54d0-4e1e-aef3-a36b72c760d5

        1 Reply Last reply
        0
        • nivocN Offline
          nivocN Offline
          nivoc
          wrote on last edited by
          #356

          In the new year I will test with "Magnesium chloride". It brings the rel humitity to exactly 32% - I will report how that goes :-)

          1 Reply Last reply
          0
          • tbowmoT Offline
            tbowmoT Offline
            tbowmo
            Admin
            wrote on last edited by tbowmo
            #357

            If the Si7021 is exposed to high humidity for a prolonged period, the RH% measurement will shift upwards, according to the datasheet, page 15, section 4.3

            Extract from Datasheet:

            4.3. Prolonged Exposure to High Humidity
            Prolonged exposure to high humidity will result in a gradual upward drift of the RH reading. The shift in sensor
            reading resulting from this drift will generally disappear slowly under normal ambient conditions. The amount of
            shift is proportional to the magnitude of relative humidity and the length of exposure. In the case of lengthy
            exposure to high humidity, some of the resulting shift may persist indefinitely under typical conditions. It is generally
            possible to substantially reverse this affect by baking the device (see Section “4.6. Bake/Hydrate Procedure” ).

            nivocN 1 Reply Last reply
            0
            • Z Offline
              Z Offline
              Zeph
              Hero Member
              wrote on last edited by Zeph
              #358

              @petoulachi said:

              BTW, what is the F() function ? instead of Serial.print("Sensebender Micro FW "); why using Serial.print(F("Sensebender Micro FW ")); ?

              It's a way to save some RAM. A simple string constant like "Hello" takes up 6 bytes of RAM and also 6 bytes of Flash (program) memory (6 bytes includes a single "hidden" byte of binary zero as an end-of-text marker). At startup and before your code executes, that bit of program memory is copied to RAM. The F() thing is a "macro" which causes the compiler save only the 6 bytes of Flash (no Ram used). The print function can (through C++ typing) fetch those 6 bytes from Flash at runtime for printing. If you have long or many text constants in your program, this can add up to some useful savings,when you only have 2K of RAM.

              See the "F() Macro" on this page: https://www.arduino.cc/en/Reference/PROGMEM

              1 Reply Last reply
              1
              • tbowmoT tbowmo

                If the Si7021 is exposed to high humidity for a prolonged period, the RH% measurement will shift upwards, according to the datasheet, page 15, section 4.3

                Extract from Datasheet:

                4.3. Prolonged Exposure to High Humidity
                Prolonged exposure to high humidity will result in a gradual upward drift of the RH reading. The shift in sensor
                reading resulting from this drift will generally disappear slowly under normal ambient conditions. The amount of
                shift is proportional to the magnitude of relative humidity and the length of exposure. In the case of lengthy
                exposure to high humidity, some of the resulting shift may persist indefinitely under typical conditions. It is generally
                possible to substantially reverse this affect by baking the device (see Section “4.6. Bake/Hydrate Procedure” ).

                nivocN Offline
                nivocN Offline
                nivoc
                wrote on last edited by
                #359

                @tbowmo Good hint. But it is happening to all three sensbeners for at least one month now. (different orders). I now got 4 more so I will test again soon.

                Over xmas I tested also with magnesium chloride hexahydrate. magnesium chloride hexahydrate produces exactly 33%rel.

                All 3 sensbenders report exactly 39% or 40%. So they are also here 7% off. But almost the same value and very reproducible - so thats great!

                upload-6b1782fc-cf3e-45b8-ad15-890b7eef84a0

                nivocN 1 Reply Last reply
                1
                • nivocN nivoc

                  @tbowmo Good hint. But it is happening to all three sensbeners for at least one month now. (different orders). I now got 4 more so I will test again soon.

                  Over xmas I tested also with magnesium chloride hexahydrate. magnesium chloride hexahydrate produces exactly 33%rel.

                  All 3 sensbenders report exactly 39% or 40%. So they are also here 7% off. But almost the same value and very reproducible - so thats great!

                  upload-6b1782fc-cf3e-45b8-ad15-890b7eef84a0

                  nivocN Offline
                  nivocN Offline
                  nivoc
                  wrote on last edited by nivoc
                  #360

                  @tbowmo I'm now almost sure that my test-setup has no faults very reproducable and read many articles - I plan to verify my findings with someone who has a professional calibrated device.

                  I saw in a museum nearby that they have professional calibrated devices :-)
                  upload-4464b273-5525-48b2-a940-10826e61f030

                  So I will test against those :-)

                  1 Reply Last reply
                  0
                  • tbowmoT Offline
                    tbowmoT Offline
                    tbowmo
                    Admin
                    wrote on last edited by
                    #361

                    @nivoc

                    Have you tried to "bake" your sensebenders, in case they have been exposed to high humidity for a longer time?

                    They are, as far as I know, produced in a part of China, which are known to have very high humidity. (been traveling there a lot some years ago)

                    1 Reply Last reply
                    0
                    • HeinzH Offline
                      HeinzH Offline
                      Heinz
                      Hero Member
                      wrote on last edited by Heinz
                      #362

                      @nivoc
                      I can confirm your results: I am running two different sensors in the same room next to each other. The sensebender is always above the dht-22. Right now it is
                      52% to 47.8% rel. humidity, while the temperature is 17.8 to 17.9 (which is nearly equal).

                      1 Reply Last reply
                      0
                      • nivocN Offline
                        nivocN Offline
                        nivoc
                        wrote on last edited by
                        #363

                        @tbowmo No I haven't baked yet. I will try that but not in the next two weeks. But I will report back.

                        1 Reply Last reply
                        0
                        • nivocN Offline
                          nivocN Offline
                          nivoc
                          wrote on last edited by
                          #364

                          But I tested tonight with distilled water - so should be exactly 100%. And again (I didn't know that thats possible) I got 104 and 107% so slightly above (still increasing). Thats great b/c that means all values are about 7-8% too high.
                          At 33% the reading was about 40%
                          At 75% the reading was about 83%
                          At 100% the reading was about 106%

                          So reading minus 8% gives a pretty accurate result. In the next few days i will test 3 more sensbender.

                          Distilled Water (100% expected):
                          upload-c0364fcd-876d-4701-a3ec-b0c55efc04a6

                          1 Reply Last reply
                          0
                          • bjacobseB Offline
                            bjacobseB Offline
                            bjacobse
                            wrote on last edited by bjacobse
                            #365

                            Read the spec for the humidity sensor used in the sensebender (Si7021 integrated humidity / temperature sensor.)
                            Precision Relative Humidity Sensor: ± 3% RH (max), 0–80% RH
                            To my reading this is actually 6% wrong reading to one of the sides - so when you are measuring yours to be 8% off, is caused that you are using the device outside of humidity range 100% and max is 80%

                            Link to spec:
                            https://www.silabs.com/Support Documents/TechnicalDocs/Si7021-A20.pdf

                            What I want you to know, if your device is reading within 6 % of the value - it's still reading correct. and when you compare ex DHT22 that device also have some reading variation error. Do not expect to get 100% perfect value for 2 devices...

                            1 Reply Last reply
                            0
                            • nivocN Offline
                              nivocN Offline
                              nivoc
                              wrote on last edited by nivoc
                              #366

                              @bjacobse Yes and no :-)

                              Yes 100% is outside of the spec.

                              And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                              Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                              M bjacobseB 2 Replies Last reply
                              0
                              • nivocN nivoc

                                @bjacobse Yes and no :-)

                                Yes 100% is outside of the spec.

                                And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                                Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                                M Offline
                                M Offline
                                mvader
                                wrote on last edited by
                                #367

                                @nivoc said:

                                @bjacobse Yes and no :-)

                                Yes 100% is outside of the spec.

                                And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                                Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                                I hope the Temp measurements are not off by that much.
                                i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                                bjacobseB nivocN 2 Replies Last reply
                                0
                                • M mvader

                                  @nivoc said:

                                  @bjacobse Yes and no :-)

                                  Yes 100% is outside of the spec.

                                  And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                                  Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                                  I hope the Temp measurements are not off by that much.
                                  i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                                  bjacobseB Offline
                                  bjacobseB Offline
                                  bjacobse
                                  wrote on last edited by
                                  #368

                                  @mvader
                                  Spec is: High Accuracy Temperature Sensor ±0.4 °C (max), –10 to 85 °C
                                  So the temp accuracy is quite good

                                  1 Reply Last reply
                                  0
                                  • nivocN nivoc

                                    @bjacobse Yes and no :-)

                                    Yes 100% is outside of the spec.

                                    And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                                    Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                                    bjacobseB Offline
                                    bjacobseB Offline
                                    bjacobse
                                    wrote on last edited by
                                    #369

                                    @nivoc
                                    Yes you are right, +-3 from your humidity solution on 33%, then the reading could be in the interval 33-3=30% or up to 33+3=36% and should not show 40%

                                    did you get same false readings from other sensebender humidity sensors?

                                    nivocN 1 Reply Last reply
                                    0
                                    • bjacobseB bjacobse

                                      @nivoc
                                      Yes you are right, +-3 from your humidity solution on 33%, then the reading could be in the interval 33-3=30% or up to 33+3=36% and should not show 40%

                                      did you get same false readings from other sensebender humidity sensors?

                                      nivocN Offline
                                      nivocN Offline
                                      nivoc
                                      wrote on last edited by
                                      #370

                                      @bjacobse Yes I have 3 sensbener running from 2 different orders. All 3 are very near by each other. So at 33% they all read something near to 40%.

                                      I have 4 more sensbenders to build. However I continue to test tonight with a Boveda-Pack that claims to bring the hum in a containter to exactly 69%. I did 4 test's so far and also used different technics. So i get more and more confident.

                                      But the very good thing is - it looks every reproducible. So the value may be off but reproducable accurate off by about 8%. When I completed my test it will be very easy to correct it within the software.

                                      1 Reply Last reply
                                      0
                                      • M mvader

                                        @nivoc said:

                                        @bjacobse Yes and no :-)

                                        Yes 100% is outside of the spec.

                                        And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                                        Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                                        I hope the Temp measurements are not off by that much.
                                        i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                                        nivocN Offline
                                        nivocN Offline
                                        nivoc
                                        wrote on last edited by
                                        #371

                                        @mvader I haven't made any very accurate temp testings. But the readings are exactly the same than other non-sensbender devices in my home. I started researching the humidity b/c I had different readings between different devices.

                                        Temp looks fine to me.

                                        1 Reply Last reply
                                        0
                                        • T Offline
                                          T Offline
                                          tcontrada
                                          wrote on last edited by
                                          #372

                                          What is the current consumption of the Sensebender during sleep and active modes?

                                          Thanks,
                                          Tony

                                          1 Reply Last reply
                                          0
                                          Reply
                                          • Reply as topic
                                          Log in to reply
                                          • Oldest to Newest
                                          • Newest to Oldest
                                          • Most Votes


                                          8

                                          Online

                                          11.7k

                                          Users

                                          11.2k

                                          Topics

                                          113.0k

                                          Posts


                                          Copyright 2019 TBD   |   Forum Guidelines   |   Privacy Policy   |   Terms of Service
                                          • Login

                                          • Don't have an account? Register

                                          • Login or register to search.
                                          • First post
                                            Last post
                                          0
                                          • MySensors
                                          • OpenHardware.io
                                          • Categories
                                          • Recent
                                          • Tags
                                          • Popular