Calibrating Humidty

Scheduled Pinned Locked Moved General Discussion
21 Posts 5 Posters 18.2k Views 2 Watching
  • Oldest to Newest
  • Newest to Oldest
  • Most Votes
Reply
  • Reply as topic
Log in to reply
This topic has been deleted. Only users with topic management privileges can see it.
  • clippermiamiC Offline
    clippermiamiC Offline
    clippermiami
    Hero Member
    wrote on last edited by
    #3

    I have some on order, I think i'm going to standardize on the DH22 for my temp/humisity sensors

    1 Reply Last reply
    0
    • YveauxY Offline
      YveauxY Offline
      Yveaux
      Mod
      wrote on last edited by
      #4

      Most 'high quality' humidity sensors come factory calibrated (e.g. Sensirion's SHTxx).
      So I doubt if calibrating them yourself will bring anything...

      http://yveaux.blogspot.nl

      clippermiamiC 1 Reply Last reply
      0
      • YveauxY Yveaux

        Most 'high quality' humidity sensors come factory calibrated (e.g. Sensirion's SHTxx).
        So I doubt if calibrating them yourself will bring anything...

        clippermiamiC Offline
        clippermiamiC Offline
        clippermiami
        Hero Member
        wrote on last edited by
        #5

        @Yveaux I did it because I wasn't sure about the quality of the mass-market DH11 sensor. I offered the tip in case anyone felt the need to confirm the units they get from China :)

        1 Reply Last reply
        0
        • hekH hek

          Thanks for the tip. You should try the same with a DHT22.

          clippermiamiC Offline
          clippermiamiC Offline
          clippermiami
          Hero Member
          wrote on last edited by clippermiami
          #6

          @hek I received my DH22s on Friday and set up another "calibration" test. i replaced the DH11 with a DH22 in the same sensor and did the "plastic bag/salt trick." After 24 hours the results were interesting.

          The DH11 registered the benchmark at 73.3, within less than 2%
          The DH22 registered the benchmark at 63.2, almost 13% low.

          I'm going to swap DH22s and see what happens. If the results are consistent with both then I will add an adjustment factor in the sketch. If they are widely different then ...

          clippermiamiC 1 Reply Last reply
          0
          • clippermiamiC clippermiami

            @hek I received my DH22s on Friday and set up another "calibration" test. i replaced the DH11 with a DH22 in the same sensor and did the "plastic bag/salt trick." After 24 hours the results were interesting.

            The DH11 registered the benchmark at 73.3, within less than 2%
            The DH22 registered the benchmark at 63.2, almost 13% low.

            I'm going to swap DH22s and see what happens. If the results are consistent with both then I will add an adjustment factor in the sketch. If they are widely different then ...

            clippermiamiC Offline
            clippermiamiC Offline
            clippermiami
            Hero Member
            wrote on last edited by
            #7

            @clippermiami Further to this I swapped the DH22 with a second unit and replenished the salt/water misture just to make sure I have a fresh start. The second DH11 reads 55% humidity, 19% below the calibration point of 75%. It appears the while they have a much narrower humidity range the DH11s are more accurate "out of the box."

            1 Reply Last reply
            0
            • YveauxY Offline
              YveauxY Offline
              Yveaux
              Mod
              wrote on last edited by
              #8

              Could you repeat the test with 2 sensors exposed to the same humidity?
              E.g. use the first DH11 which gave results very close to 75% together with the DH22 which was very off?
              Then you also know the repeatability of your measurement.

              http://yveaux.blogspot.nl

              clippermiamiC 1 Reply Last reply
              0
              • hekH Online
                hekH Online
                hek
                Admin
                wrote on last edited by
                #9

                Just got some samples of the SHT21 from Sensirion. Damn it's small.

                Just have to figure out how to surface mount it on something.

                YveauxY 1 Reply Last reply
                0
                • YveauxY Yveaux

                  Could you repeat the test with 2 sensors exposed to the same humidity?
                  E.g. use the first DH11 which gave results very close to 75% together with the DH22 which was very off?
                  Then you also know the repeatability of your measurement.

                  clippermiamiC Offline
                  clippermiamiC Offline
                  clippermiami
                  Hero Member
                  wrote on last edited by
                  #10

                  @Yveaux I want to but i do't have the tools to do it right now.

                  BulldogLowellB 1 Reply Last reply
                  0
                  • clippermiamiC clippermiami

                    @Yveaux I want to but i do't have the tools to do it right now.

                    BulldogLowellB Offline
                    BulldogLowellB Offline
                    BulldogLowell
                    Contest Winner
                    wrote on last edited by
                    #11

                    @clippermiami said:

                    @Yveaux I want to but i do't have the tools to do it right now.

                    How do they check out at the high end of the scale (i.e. 99-100% RH in a sealed polybag with a damp towel)?

                    I tested my Aeons in a temperature/humidity controlled chamber in our lab and got a spread of 10%RH points on two versus known (tested at 72F and 50%RH). I thought that was pretty poor.

                    I have not run any tests with my DHT22' but I may based on this.

                    Thanks for the info, very helpful.

                    Jim

                    clippermiamiC 1 Reply Last reply
                    0
                    • hekH hek

                      Just got some samples of the SHT21 from Sensirion. Damn it's small.

                      Just have to figure out how to surface mount it on something.

                      YveauxY Offline
                      YveauxY Offline
                      Yveaux
                      Mod
                      wrote on last edited by
                      #12

                      @hek How are your eyes ??? ;-)
                      I once tried soldering wires to one, but it lost a 'leg' in the process... $15 down the drain...
                      Better use a small breakout to solder them on!

                      http://yveaux.blogspot.nl

                      1 Reply Last reply
                      0
                      • BulldogLowellB BulldogLowell

                        @clippermiami said:

                        @Yveaux I want to but i do't have the tools to do it right now.

                        How do they check out at the high end of the scale (i.e. 99-100% RH in a sealed polybag with a damp towel)?

                        I tested my Aeons in a temperature/humidity controlled chamber in our lab and got a spread of 10%RH points on two versus known (tested at 72F and 50%RH). I thought that was pretty poor.

                        I have not run any tests with my DHT22' but I may based on this.

                        Thanks for the info, very helpful.

                        Jim

                        clippermiamiC Offline
                        clippermiamiC Offline
                        clippermiami
                        Hero Member
                        wrote on last edited by clippermiami
                        #13

                        @BulldogLowell The results with the DH22 were surprisingly disappointing, I thought they would have better performance. I tried two of them without any correction factor in the sketch and in my 75% humidity test they never got close. They were however consistent so applying a (rather large) correction factor in the code would bring them into "zero" at 75%.

                        In my 75% "chamber"
                        DH22 # 1 reported 55.8
                        DH22 # 2 reported 55.4

                        so they are pretty close to each other. A uniform adjustment of say +20 would bring them very close. However in free air the results are different

                        My two Aeon 4-in-1 report reading of 55% and 43% so they are pretty far apart as is. Tamiami Airport (General Aviation) is about 5 miles SW of us and it reports the still air outside humidity at 55% both yesterday and today --- so allowing for differences in inside (A/C) vs outside humidity I'd guess the actual inside at somewhere around 40-45%

                        In free air inside:
                        DH22 # 1 reported 39.0%
                        DH22 # 2 reported 38.6%

                        so they may be fairly accurate at normal room humidity levels. But applying a +20 correction factor to make them read correctly in the testbed would trow the free air levels way off.

                        I've put the DH11 back in the testbed to see what happens.

                        Of course at the end of the day I just don't have the benchmark tools to do a proper comparison at home so this is all pretty loose.

                        John

                        BulldogLowellB YveauxY 2 Replies Last reply
                        0
                        • clippermiamiC clippermiami

                          @BulldogLowell The results with the DH22 were surprisingly disappointing, I thought they would have better performance. I tried two of them without any correction factor in the sketch and in my 75% humidity test they never got close. They were however consistent so applying a (rather large) correction factor in the code would bring them into "zero" at 75%.

                          In my 75% "chamber"
                          DH22 # 1 reported 55.8
                          DH22 # 2 reported 55.4

                          so they are pretty close to each other. A uniform adjustment of say +20 would bring them very close. However in free air the results are different

                          My two Aeon 4-in-1 report reading of 55% and 43% so they are pretty far apart as is. Tamiami Airport (General Aviation) is about 5 miles SW of us and it reports the still air outside humidity at 55% both yesterday and today --- so allowing for differences in inside (A/C) vs outside humidity I'd guess the actual inside at somewhere around 40-45%

                          In free air inside:
                          DH22 # 1 reported 39.0%
                          DH22 # 2 reported 38.6%

                          so they may be fairly accurate at normal room humidity levels. But applying a +20 correction factor to make them read correctly in the testbed would trow the free air levels way off.

                          I've put the DH11 back in the testbed to see what happens.

                          Of course at the end of the day I just don't have the benchmark tools to do a proper comparison at home so this is all pretty loose.

                          John

                          BulldogLowellB Offline
                          BulldogLowellB Offline
                          BulldogLowell
                          Contest Winner
                          wrote on last edited by
                          #14

                          @clippermiami

                          wow, I am surprised at the poor results. :(

                          repeatable is important, but precision needs accuracy, too

                          thanks for this; it is making me re-think some of what I'm doing with sensors.

                          clippermiamiC 1 Reply Last reply
                          0
                          • BulldogLowellB BulldogLowell

                            @clippermiami

                            wow, I am surprised at the poor results. :(

                            repeatable is important, but precision needs accuracy, too

                            thanks for this; it is making me re-think some of what I'm doing with sensors.

                            clippermiamiC Offline
                            clippermiamiC Offline
                            clippermiami
                            Hero Member
                            wrote on last edited by
                            #15

                            @BulldogLowell I think it would be good to have a few more folks to do some tests and report the results. I think I will start from scratch on the data as well just to verify my results

                            BulldogLowellB 1 Reply Last reply
                            0
                            • clippermiamiC clippermiami

                              @BulldogLowell I think it would be good to have a few more folks to do some tests and report the results. I think I will start from scratch on the data as well just to verify my results

                              BulldogLowellB Offline
                              BulldogLowellB Offline
                              BulldogLowell
                              Contest Winner
                              wrote on last edited by
                              #16

                              @clippermiami said:

                              @BulldogLowell I think it would be good to have a few more folks to do some tests and report the results. I think I will start from scratch on the data as well just to verify my results

                              I could test a couple too, if you want. I'll get them back in the humidity chamber in our lab.

                              I am still thinking about the salt test and wondering how the temperature may play a role (since RH is dependent on temperature). I can't seem to get myself to the point where I understand that the salt/water paste will get any (albeit tiny) air atmosphere to 75%RH independent of temperature, unless that is a "generally at room temperature" qualification.

                              I would like to calculate an offset. I know it is 'correct' at a saturation humidity so I imagine the offset will be some kind of formula off of the results of two or three or more known values.

                              Actually, that would be a great arduino project in itself... develop an offset curve with 3-4 calibration points.

                              1 Reply Last reply
                              0
                              • clippermiamiC clippermiami

                                @BulldogLowell The results with the DH22 were surprisingly disappointing, I thought they would have better performance. I tried two of them without any correction factor in the sketch and in my 75% humidity test they never got close. They were however consistent so applying a (rather large) correction factor in the code would bring them into "zero" at 75%.

                                In my 75% "chamber"
                                DH22 # 1 reported 55.8
                                DH22 # 2 reported 55.4

                                so they are pretty close to each other. A uniform adjustment of say +20 would bring them very close. However in free air the results are different

                                My two Aeon 4-in-1 report reading of 55% and 43% so they are pretty far apart as is. Tamiami Airport (General Aviation) is about 5 miles SW of us and it reports the still air outside humidity at 55% both yesterday and today --- so allowing for differences in inside (A/C) vs outside humidity I'd guess the actual inside at somewhere around 40-45%

                                In free air inside:
                                DH22 # 1 reported 39.0%
                                DH22 # 2 reported 38.6%

                                so they may be fairly accurate at normal room humidity levels. But applying a +20 correction factor to make them read correctly in the testbed would trow the free air levels way off.

                                I've put the DH11 back in the testbed to see what happens.

                                Of course at the end of the day I just don't have the benchmark tools to do a proper comparison at home so this is all pretty loose.

                                John

                                YveauxY Offline
                                YveauxY Offline
                                Yveaux
                                Mod
                                wrote on last edited by
                                #17

                                @clippermiami So the 'inside' reading seems to be closer to the expected value than the 'salt chamber' value. This suggests using just an offset for calibration is not sufficient, but at least gain/offset and possibly an interpolation table are required to get some accuracy...

                                http://yveaux.blogspot.nl

                                clippermiamiC 1 Reply Last reply
                                0
                                • YveauxY Yveaux

                                  @clippermiami So the 'inside' reading seems to be closer to the expected value than the 'salt chamber' value. This suggests using just an offset for calibration is not sufficient, but at least gain/offset and possibly an interpolation table are required to get some accuracy...

                                  clippermiamiC Offline
                                  clippermiamiC Offline
                                  clippermiami
                                  Hero Member
                                  wrote on last edited by
                                  #18

                                  @Yveaux The data sheet for the DHT22 (the DHT11 is in Chinese) says that the sensor is fully calibrated and has in internal lookup table to adjust the values across the entire temperature range. S from that one would assume that the output of the device is linear.

                                  I am just finishing up a long (48 hour) test using the salt-water-plastic-bag test jig using the DHT11. Right now it shows 76% humidity and that's after I recompiled the sketch with a a +2% adjustment. In this case +1% adjustment would have put it right on the money at 75%. The 76% value has been stable for the past 24 hours so I am satisfied that the reading is correct and would plan to use a 1% adjustment.

                                  This morning I'm going to take it out of the test jig and let it settle to the ambient room humidity and try to judge where I am at. We shall see. But at least i know the DHT11 can correctly hit the expected test value. I may just settle on the DHT11 for my sensors.

                                  I'm working on two 3V battery operated Minis' with Hum/Temp/Barometric/Light sensors to go outside and i'm gathering parts for 8-9 battery operated Motion/Humidity/Temp/Light Level sensors for indoor use. All will be powered by CR123A 3V Lithium batts.

                                  YveauxY 1 Reply Last reply
                                  0
                                  • clippermiamiC clippermiami

                                    @Yveaux The data sheet for the DHT22 (the DHT11 is in Chinese) says that the sensor is fully calibrated and has in internal lookup table to adjust the values across the entire temperature range. S from that one would assume that the output of the device is linear.

                                    I am just finishing up a long (48 hour) test using the salt-water-plastic-bag test jig using the DHT11. Right now it shows 76% humidity and that's after I recompiled the sketch with a a +2% adjustment. In this case +1% adjustment would have put it right on the money at 75%. The 76% value has been stable for the past 24 hours so I am satisfied that the reading is correct and would plan to use a 1% adjustment.

                                    This morning I'm going to take it out of the test jig and let it settle to the ambient room humidity and try to judge where I am at. We shall see. But at least i know the DHT11 can correctly hit the expected test value. I may just settle on the DHT11 for my sensors.

                                    I'm working on two 3V battery operated Minis' with Hum/Temp/Barometric/Light sensors to go outside and i'm gathering parts for 8-9 battery operated Motion/Humidity/Temp/Light Level sensors for indoor use. All will be powered by CR123A 3V Lithium batts.

                                    YveauxY Offline
                                    YveauxY Offline
                                    Yveaux
                                    Mod
                                    wrote on last edited by
                                    #19

                                    @clippermiami said:

                                    from that one would assume that the output of the device is linear.

                                    From that one would assume the output of the device is calibrated, I would say ;-)

                                    http://yveaux.blogspot.nl

                                    clippermiamiC 1 Reply Last reply
                                    0
                                    • YveauxY Yveaux

                                      @clippermiami said:

                                      from that one would assume that the output of the device is linear.

                                      From that one would assume the output of the device is calibrated, I would say ;-)

                                      clippermiamiC Offline
                                      clippermiamiC Offline
                                      clippermiami
                                      Hero Member
                                      wrote on last edited by
                                      #20

                                      @Yveaux Good point

                                      1 Reply Last reply
                                      0
                                      • epierreE Offline
                                        epierreE Offline
                                        epierre
                                        Hero Member
                                        wrote on last edited by
                                        #21

                                        Hello,

                                        I have tested the DHT11 for one day with a THGR810 from Oregon (assuming it is pretty calibrated). They were close together.

                                        There are some time the DHT11 moved but not the Oregon. Generally it was 1% below so pretty good.

                                        On Temperature there is more gap, it varies from 2°C+/-1°C below, it is more puzzling here.

                                        I have not yet DHT22 to do this check-up.

                                        Would one day adjustment values sent by the gateway to the sensors ;-)

                                        z-wave - Vera -> Domoticz
                                        rfx - Domoticz <- MyDomoAtHome <- Imperihome
                                        mysensors -> mysensors-gw -> Domoticz

                                        1 Reply Last reply
                                        0
                                        Reply
                                        • Reply as topic
                                        Log in to reply
                                        • Oldest to Newest
                                        • Newest to Oldest
                                        • Most Votes


                                        14

                                        Online