Skip to content
  • MySensors
  • OpenHardware.io
  • Categories
  • Recent
  • Tags
  • Popular
Skins
  • Light
  • Brite
  • Cerulean
  • Cosmo
  • Flatly
  • Journal
  • Litera
  • Lumen
  • Lux
  • Materia
  • Minty
  • Morph
  • Pulse
  • Sandstone
  • Simplex
  • Sketchy
  • Spacelab
  • United
  • Yeti
  • Zephyr
  • Dark
  • Cyborg
  • Darkly
  • Quartz
  • Slate
  • Solar
  • Superhero
  • Vapor

  • Default (No Skin)
  • No Skin
Collapse
Brand Logo
  1. Home
  2. Announcements
  3. Sensebender Micro

Sensebender Micro

Scheduled Pinned Locked Moved Announcements
584 Posts 84 Posters 548.2k Views 35 Watching
  • Oldest to Newest
  • Newest to Oldest
  • Most Votes
Reply
  • Reply as topic
Log in to reply
This topic has been deleted. Only users with topic management privileges can see it.
  • nivocN nivoc

    @tbowmo Good hint. But it is happening to all three sensbeners for at least one month now. (different orders). I now got 4 more so I will test again soon.

    Over xmas I tested also with magnesium chloride hexahydrate. magnesium chloride hexahydrate produces exactly 33%rel.

    All 3 sensbenders report exactly 39% or 40%. So they are also here 7% off. But almost the same value and very reproducible - so thats great!

    upload-6b1782fc-cf3e-45b8-ad15-890b7eef84a0

    nivocN Offline
    nivocN Offline
    nivoc
    wrote on last edited by nivoc
    #360

    @tbowmo I'm now almost sure that my test-setup has no faults very reproducable and read many articles - I plan to verify my findings with someone who has a professional calibrated device.

    I saw in a museum nearby that they have professional calibrated devices :-)
    upload-4464b273-5525-48b2-a940-10826e61f030

    So I will test against those :-)

    1 Reply Last reply
    0
    • T Offline
      T Offline
      tbowmo
      Admin
      wrote on last edited by
      #361

      @nivoc

      Have you tried to "bake" your sensebenders, in case they have been exposed to high humidity for a longer time?

      They are, as far as I know, produced in a part of China, which are known to have very high humidity. (been traveling there a lot some years ago)

      1 Reply Last reply
      0
      • HeinzH Offline
        HeinzH Offline
        Heinz
        Hero Member
        wrote on last edited by Heinz
        #362

        @nivoc
        I can confirm your results: I am running two different sensors in the same room next to each other. The sensebender is always above the dht-22. Right now it is
        52% to 47.8% rel. humidity, while the temperature is 17.8 to 17.9 (which is nearly equal).

        1 Reply Last reply
        0
        • nivocN Offline
          nivocN Offline
          nivoc
          wrote on last edited by
          #363

          @tbowmo No I haven't baked yet. I will try that but not in the next two weeks. But I will report back.

          1 Reply Last reply
          0
          • nivocN Offline
            nivocN Offline
            nivoc
            wrote on last edited by
            #364

            But I tested tonight with distilled water - so should be exactly 100%. And again (I didn't know that thats possible) I got 104 and 107% so slightly above (still increasing). Thats great b/c that means all values are about 7-8% too high.
            At 33% the reading was about 40%
            At 75% the reading was about 83%
            At 100% the reading was about 106%

            So reading minus 8% gives a pretty accurate result. In the next few days i will test 3 more sensbender.

            Distilled Water (100% expected):
            upload-c0364fcd-876d-4701-a3ec-b0c55efc04a6

            1 Reply Last reply
            0
            • B Offline
              B Offline
              bjacobse
              wrote on last edited by bjacobse
              #365

              Read the spec for the humidity sensor used in the sensebender (Si7021 integrated humidity / temperature sensor.)
              Precision Relative Humidity Sensor: ± 3% RH (max), 0–80% RH
              To my reading this is actually 6% wrong reading to one of the sides - so when you are measuring yours to be 8% off, is caused that you are using the device outside of humidity range 100% and max is 80%

              Link to spec:
              https://www.silabs.com/Support Documents/TechnicalDocs/Si7021-A20.pdf

              What I want you to know, if your device is reading within 6 % of the value - it's still reading correct. and when you compare ex DHT22 that device also have some reading variation error. Do not expect to get 100% perfect value for 2 devices...

              1 Reply Last reply
              0
              • nivocN Offline
                nivocN Offline
                nivoc
                wrote on last edited by nivoc
                #366

                @bjacobse Yes and no :-)

                Yes 100% is outside of the spec.

                And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                M B 2 Replies Last reply
                0
                • nivocN nivoc

                  @bjacobse Yes and no :-)

                  Yes 100% is outside of the spec.

                  And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                  Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                  M Offline
                  M Offline
                  mvader
                  wrote on last edited by
                  #367

                  @nivoc said:

                  @bjacobse Yes and no :-)

                  Yes 100% is outside of the spec.

                  And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                  Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                  I hope the Temp measurements are not off by that much.
                  i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                  B nivocN 2 Replies Last reply
                  0
                  • M mvader

                    @nivoc said:

                    @bjacobse Yes and no :-)

                    Yes 100% is outside of the spec.

                    And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                    Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                    I hope the Temp measurements are not off by that much.
                    i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                    B Offline
                    B Offline
                    bjacobse
                    wrote on last edited by
                    #368

                    @mvader
                    Spec is: High Accuracy Temperature Sensor ±0.4 °C (max), –10 to 85 °C
                    So the temp accuracy is quite good

                    1 Reply Last reply
                    0
                    • nivocN nivoc

                      @bjacobse Yes and no :-)

                      Yes 100% is outside of the spec.

                      And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                      Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                      B Offline
                      B Offline
                      bjacobse
                      wrote on last edited by
                      #369

                      @nivoc
                      Yes you are right, +-3 from your humidity solution on 33%, then the reading could be in the interval 33-3=30% or up to 33+3=36% and should not show 40%

                      did you get same false readings from other sensebender humidity sensors?

                      nivocN 1 Reply Last reply
                      0
                      • B bjacobse

                        @nivoc
                        Yes you are right, +-3 from your humidity solution on 33%, then the reading could be in the interval 33-3=30% or up to 33+3=36% and should not show 40%

                        did you get same false readings from other sensebender humidity sensors?

                        nivocN Offline
                        nivocN Offline
                        nivoc
                        wrote on last edited by
                        #370

                        @bjacobse Yes I have 3 sensbener running from 2 different orders. All 3 are very near by each other. So at 33% they all read something near to 40%.

                        I have 4 more sensbenders to build. However I continue to test tonight with a Boveda-Pack that claims to bring the hum in a containter to exactly 69%. I did 4 test's so far and also used different technics. So i get more and more confident.

                        But the very good thing is - it looks every reproducible. So the value may be off but reproducable accurate off by about 8%. When I completed my test it will be very easy to correct it within the software.

                        1 Reply Last reply
                        0
                        • M mvader

                          @nivoc said:

                          @bjacobse Yes and no :-)

                          Yes 100% is outside of the spec.

                          And no. 33% with Magnesium Chloride Hexahydrate is inside the spec and I get 40% instead of 33-34%.

                          Max off should be 3% above or? Means 37% - but fair enough. Should be ok for household use.

                          I hope the Temp measurements are not off by that much.
                          i don't need .00001 accuracy or anything. but i feel a big difference between 70 and 72 in my house.

                          nivocN Offline
                          nivocN Offline
                          nivoc
                          wrote on last edited by
                          #371

                          @mvader I haven't made any very accurate temp testings. But the readings are exactly the same than other non-sensbender devices in my home. I started researching the humidity b/c I had different readings between different devices.

                          Temp looks fine to me.

                          1 Reply Last reply
                          0
                          • T Offline
                            T Offline
                            tcontrada
                            wrote on last edited by
                            #372

                            What is the current consumption of the Sensebender during sleep and active modes?

                            Thanks,
                            Tony

                            1 Reply Last reply
                            0
                            • T Offline
                              T Offline
                              tbowmo
                              Admin
                              wrote on last edited by
                              #373

                              @tcontrada

                              there are some measurements in the old design thread here http://forum.mysensors.org/topic/510/minimal-design-thoughts

                              1 Reply Last reply
                              1
                              • N Offline
                                N Offline
                                nikos1671
                                wrote on last edited by
                                #374

                                Hi. What case you all using for the sensebender ?Any good idea.

                                1 Reply Last reply
                                0
                                • F Offline
                                  F Offline
                                  filipq
                                  wrote on last edited by
                                  #375

                                  3_1456326631089_IMG_20160224_150344.jpg 2_1456326631089_IMG_20160224_150314.jpg 1_1456326631089_IMG_20160224_150303.jpg 0_1456326631088_IMG_20160224_150237.jpg

                                  1 Reply Last reply
                                  1
                                  • T Offline
                                    T Offline
                                    tbowmo
                                    Admin
                                    wrote on last edited by
                                    #376

                                    @filipq

                                    Where did you find that box?

                                    1 Reply Last reply
                                    0
                                    • F Offline
                                      F Offline
                                      filipq
                                      wrote on last edited by
                                      #377

                                      @tbowmo

                                      Got it from http://cpc.farnell.com/1/1/57932-evatron-pp42m-case-sensor-ivory-71x71x27mm.html.
                                      Not the cheapest, but definitely worth the money as it is very good quality. There are other colours available (white and black) https://www.jprelec.co.uk/store.asp/c=552/Gas-Sensor-Case

                                      1 Reply Last reply
                                      0
                                      • S Offline
                                        S Offline
                                        scalz
                                        Hardware Contributor
                                        wrote on last edited by
                                        #378

                                        @tbowmo: I have some of these too, it's nice box. i got these from tme. cheaper :)
                                        http://www.tme.eu/fr/details/box-sens-white/boitiers-dappar-dalarme-et-de-capt/supertronic/

                                        1 Reply Last reply
                                        0
                                        • A Offline
                                          A Offline
                                          ahmedadelhosni
                                          wrote on last edited by
                                          #379

                                          Hello @tbowmo

                                          Doesn't the atmega328 has only 2 pins which are D2 and D3 from external interrupts ?
                                          I am wondering why do we waste the usage of D2 since the IRQ pin is not used in the library ?

                                          A 1 Reply Last reply
                                          0

                                          Hello! It looks like you're interested in this conversation, but you don't have an account yet.

                                          Getting fed up of having to scroll through the same posts each visit? When you register for an account, you'll always come back to exactly where you were before, and choose to be notified of new replies (either via email, or push notification). You'll also be able to save bookmarks and upvote posts to show your appreciation to other community members.

                                          With your input, this post could be even better 💗

                                          Register Login
                                          Reply
                                          • Reply as topic
                                          Log in to reply
                                          • Oldest to Newest
                                          • Newest to Oldest
                                          • Most Votes


                                          18

                                          Online

                                          12.0k

                                          Users

                                          11.2k

                                          Topics

                                          113.4k

                                          Posts


                                          Copyright 2025 TBD   |   Forum Guidelines   |   Privacy Policy   |   Terms of Service
                                          • Login

                                          • Don't have an account? Register

                                          • Login or register to search.
                                          • First post
                                            Last post
                                          0
                                          • MySensors
                                          • OpenHardware.io
                                          • Categories
                                          • Recent
                                          • Tags
                                          • Popular